What is TK3 calibrator?

The TK3-2 Calibration Instrument simulates shaft vibration and position. It verifies the monitor readings operating condition as well as the proximity transducer system condition. A properly calibrated system means accurate measurements of the monitor and transducer inputs.

Working:

A spindle micrometer is used to check the calibration monitoring of the transducer system and position. A TK3-2 mounting probe adapter holds the probe while the target is moved in calibrated increments to or from the tip of the probe. The Proximity sensor output is recorded using a voltmeter.

Using the motor-driven wobble plate, vibration monitors are calibrated. To hold the proximity probe in place, a swing-arm holder is located above the wobble plate.

The holder and the probe are adjusted to a position where the mechanical vibration desired is found. The vibration is then measured using the unit’s dial micrometer (no oscilloscope is required).

The reading of a vibration monitor can then be compared with the known input of the mechanical vibration signal. The TK3-2 mechanical vibration signal can range from peak-to-peak 50 to 254 μm (2 to 10 miles).

Specification:

Power
Electric: 95 - 125 Vac, 50/60 Hz, 1A minimum.
190 - 250 Vac, 50/60 Hz, 1A minimum.

Air: 90 psi (6.2 bar) maximum.
Wobulator Range
Vibration Amplitude

Range: 50 μm to 254 μm (2 to 10 mils) peak-to-peak.

Maximum Speed

Electric: 0 to 10,000 cpm ±2,500 cpm.
Air: 0 to 5,000 cpm ±1,000 cpm